Hi all,

Is anyone familar with using two (or more) internal standards (ISTD), more specifically, how can this information be added into the sequence table for report calulations?

At present I have single ISTD working well, but can only enter one ISTD concentration (per sample) into the sequence table ISTD column for the integration report to use.

Thanks.

I tend to use batch reprocessing, so things may be different, but here's a thought:

If your samples were all treated the same, and therefore contain the same amount of internal standard mix, you can set up all the internal standards in your calibration table, then go to Report - Specify report, and in "Calculation Factors", specify "From sample defaults below" instead of "from data file". You should now have a table below this box with a line for each internal standard, where you can enter different amounts for each. Good luck!
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry