DJ » Mon Mar 22, 2010 9:18 am
[quote="immergruen" DJ: Yes, well, the aqueous phase is saturated with octanol by adding 10% of octanol to the HEPES buffer, stirring it over night, putting it into a separation funnel for four hours and then using the lower layer as mobile phase. The assay is used for measuring LogD by retention time / capacity factor of the sample compared to a set of standard substances.[/quote]
I have a little bit oxperience with logD by hplc. By your inclusion of HEPES, I assume you are interested in logp in the physiological pH range. Many (but not all) basic compounds will be ionized at pH 7, however, some (imidazole-like basic groups) are not. This means long long retention times if done isocratically.
An approach I used successfully (and have -never- added octanol to my mobile phase) is to perform several isocratic runs. Plot log k vs % organic. The y intercept from this regression is logKw, the theoretical retention factor under neat buffer conditions. The other regression coefficient is S (slope obtained from linear regression.
This means a lot of runs to determine a logKw (approx related to log P), and if your already messing with sticky octanol, might as well do the conventional shake flask, eh? :_
Another way to estimate logKw and S is from retention data of two gradient runs (change tg or sub MeCN for MeOH). From this, you can approximate obtain Kw and S numerically.
With these two properties on hand, it is theoretically possible to predict retention times for ANY gradient program.