typical salt concentration in sample for SEC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
hi.. i have a basic question..
i wish to know the typically used salt concentration in the sample that is to be injected on the TOSOH SEC silica based column... should my sample be in higher salt say 100 mM salt or in lower salt.. 10 mM salt.. and why?

thanks
Explore!

The best approach is to inject the sample with the same salt concentration and buffer pH as the mobile phase. The reason is that these mobile phase conditions have been set up to be favorable for SEC, avoiding silanol interactions and similar secondary interactions. If you inject the sample in something else, there is a high chance that you will create additional interactions, and that this will influence the retention (or better: the desired lack of retention).
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry