Posted: Fri Mar 12, 2010 11:27 am
by Nela
Hi,

I have a problem with the Agilent Chemstation. Everytime I try to start a measurement the software completes the run after 1 second. There is no error message in the logbook, and pumps and injector don't have any stop time. The run is completed as usual, but right after 1 second!

Whyyyy???????????????????
It must be some stupid small mistake in the software...

Posted: Fri Mar 12, 2010 1:09 pm
by Schmitty
I haven't used LC Chemstaion in quite a few years, but I am familiar with the GC/MS Chemstation. Usually if something like that happens, "Data Analysis" is the only item selected when a run is started. When the software finds no data to process (since you haven't collected it yet, theoreticaly), it just stops.

Best of luck

Posted: Fri Mar 12, 2010 2:10 pm
by Garlarg
If you measure a single sample then be sure you checked the "Data Acquisition" and "Stadard data analysis" in the Method/Run Time Checklist menu.
If you measure sequence then check the sequence/sequence parameters menu. The "Part of methods to run" field must be "According to runtime checklist", or maybe "Acquisition only" but not "Reprocess only".

Posted: Wed Mar 17, 2010 4:28 pm
by Earl C
Is your drive full? If there is not enough memory the chemstation will only collect a portion of the run or stop it short and sometimes won't give an error.

Posted: Tue Mar 23, 2010 3:56 pm
by mbicking
Is this LC or GC? Check all method settings. In LC, each module has a Stop Time, which should follow the pump. If the pump Stop Time is a small value, then the system will stop at that point.