Essentially my question is:
Can you use a C18 RP column to remove SDS from a (peptide) sample?
I have a small peptide (50 ng of 1 kDa) which is in 100 ul of buffer containing approximately 0.05 % SDS and 0.5 M Tris. I wish to clean up the peptide prior to MALDI-MS analysis.
I have been trying to clean the peptide using standard C18 disposable columns but I am concerned the SDS is not sufficiently being removed. I have been led to believe that SDS will irreversibly bind to the C18 phase and have been using a 50 mg C18 column which should have more than enough binding capacity for the peptide and the SDS.
I have acidified my sample with TFA and eluted from the column using a stepwise gradient of acetonitrile. I have checked and I am managing to essentially recover all my peptide from the column eluting with 40 % ACN. However, my peptide does not seem to want to fly in MALDI-MS (it normally does fly).
I am concerned I have not removed all the SDS from my sample. Will all the SDS bind to the C18 (assuming sufficient capacity) or will some come through? Will SDS also be eluted by ACN?
I am working on removing the detergent by other means before the C18 column but if anyone knows how the C18 is likely performing then I would be very grateful to hear!
We are also interested in buying a C18 RP column to perform seperation but are concerned the SDS/Triton X-100 will permanently damage the column.
Yours hopefully
Adrian.