what is thereasons for different peak heights with sameconc

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
hi there is an observation was made by me during my analysis i.e my standard peak height is 500AU and sample peak height is 800AU but both have same area it is in Assay By HPLC Usp method
1.both preparations are made at same concentrations i.e 0.5mg/ml
2.HPLC system is Shimadzu Prominence UFLC with LC-Solutions database software
3.it was observed only two times but i am analysing daily the samples with same standard and same analyticalcolumn (RP18e,250x4.5mm,3µm) and also more times in same instrument

Let the speculation begin !

Two possibilities: the lower peak is wider than the higher one (for whatever reason), or 300 MAU of inconsistent baseline drift.

Peter
Peter Apps

What are the respective retention times?
What will happen if you repeat the injections?

Best Regards
Learn Innovate and Share

Dancho Dikov

That´s it: The peaks are not at the same retention time.
4 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry