Analysis problem with UV
Posted: Tue Mar 09, 2010 10:38 pm
Hello guys
I have a problem with my analysis- Thought maybe someone has a solution
I am trying to get a concentration of the Vitamin B12 with UV - Vis spectrophotometer in a finished product. So, I ran the UV spectra of the B12 and came up with 361 and 550 nm wavelengths.
Here is the problem
The finished product showed 4 times higher B12 content than it should.
I have a perfect linear curve based on 6 point standard calibration.
I made my standards in deionized water.
Then I thought-
I'll get the placebo and run that without any B12.
Here was the mystery problem.
The Placebo had about same concentration B12!! . It was about 4 times high as well. ( which it shouldn't be there)
I ran the placebo in 3ml/10 ml water dilution.
( I wish I could post the spectra for you- but I couldn't)
The spectra ran from 190 nm to 700 nm and it took a big curve right about 380 nm to all the way to the left of the spectrum.
The placebo has some Riboflavin in it which looks orange- The final product is maroon red
here is what I did to test myself
The UV that I use is Shimadzu 1700 series-
I ran the UV on two different modes
Mode 1 is Spectrum
I ran the analysis on this mode- Got the calibration points and Absorbances and the curve looked perfect- Got the Absorbance of the sample and (A-Intercept/Slope) gave me the same answer as what I did on Mode 3 which is Quantitation
Same thing again
I constructed a calibration curve and ran the sample and the placebo
and both gave me four times higher than expected result.
I'm thinking somwhere I have interferring wavelengths-
Anyway- I'm left with reading the manual to see whether I can do something else..
Here is the problem
Then I thought-
Mode 1 is Spectrum
Same thing again