ok to open drain valve with 2000 psi pump pressure?
Posted: Tue Mar 09, 2010 6:53 pm
Hi everyone,
When I was first trained in HPLC over the summer (by fellow students), I was told not to open the drain valve or loosen any fittings until the pressure had subsided to under 10 bar. However, I don't think any of us fully understand the reason for this (including whether it relates to the column, the pump, the lines, or any combo thereof), enough to be sure of the answer to the following question.
If the column is not connected, and if in fact the pumps are not even connected to the autosampler, is it still advisable to not open the drain valve unless the pressure in the pump has subsided to atmospheric levels?
The reason I ask is that I am currently doing a pressure decay test. I have now fixed the A pump to the point that it is holding pressure nomally, but there is still a leak in the pressure of the B pump. So, say I do this test and the pressure in A is 4000 psi but the pressure in B falls almost immediately, then I remove a check valve to sonicate or replace it. I must prime the pump again, which requires opening the drain valve, but the pressure in A remains high. Can I open the drain valve?
If not, is there a convenient and safe way to remove pressure from pump A?
Chris
Chris