variable retention time problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Image

peaks in red box show variable retention time. any thought? I equalibrium column for at least 3 column volume.
any comments? thanks a lot.
plus,
there is no problem in isocratic elution.
so, what is the possible source of problem?
air bubble? pump malfunction?
Image

peaks in red box show variable retention time. any thought? I equalibrium column for at least 3 column volume.
any comments? thanks a lot.

With any gradient program, if I had reproducibility issues, I'd increase the equilibration time before the next injection, then ask myself: did things get better, worse, or the same? And then go from there. You may also want to run a standard mixture several times to check out gradient performance of your instrument.

Another good thing is to decrease increase injection volume and ask yourself if peak shape got better, worse, or the same. Very valuable.

Can you leave more time for equilibrium.
Also check the pumps precision.

I'd vote for a gradient proportioning problem. This is almost identical to a problem described in one of John Dolan's ""LC Troubleshooting" columns:
LC-GC 14(4) 294-299 (1996)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

i do agree that it is gradient proportion problem since i run 3.2 bed volume for equilibrium.
but I cannot find it literature you listed.
could you give me a link ?
I'd vote for a gradient proportioning problem. This is almost identical to a problem described in one of John Dolan's ""LC Troubleshooting" columns:
LC-GC 14(4) 294-299 (1996)

e-mail me with your (physical) mailing address and I'll send you a CD that has that article (actually, all of Dolan's LC Troubleshooting articles).

For anyone else who wants one, I can only send them to mailing addresses in North America. :cry:
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
7 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Majestic-12 [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry