Something quite strange has happened in my Dionex Ultimate 3000 HPLC.
The stainless steel filter frit that is placed in the outlet unit of the pump has vanished, only the plastic washer that holds it has remained!
My eluents are:
A: 10 mM KH2PO4 ph= 2.8
B: 500 mM KH2PO4 800 mM KCl ph= 2.8
Column: Whatman Partisil 10 SAX
Eluents are at ambient temperature
I guess it was a 0.5 uM frit (because of the colour of the washer), which is NOT the recommended one for eluents of pH<4, nor for my tipe of pump. (It should have been 10 uM)
I found tiny metallic fragmens in the outlet unit, also when opening the inlet fitting of the column.
The column had a void, but I don´t know whether it has been caused by the particulates. (Whatman technician told me that packing material may have been dissolved)
Dionex technician advised me to perform a special wash to remove all the fragments before installing a new column.
The wash is HNO3 3N, which is quite strong.
Anyone suffered a similar problem?
Do you think that the acid wash is adequate?
Column void can be caused by particulates entering the column?
Thank you very much for your time, any ideas will be very welcomed!!!