Trace DSQII GC-MS detector gain calibration failed

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi

I was so excited to run samples today, after cleaning the ion source, lenses and all else that needed cleaning last week.

When I did a full tune this morning the detector gain calibraiton failed!

Any ideas why this happend??

SM

We had this problem last month

The engineer found that the settings of the lens 1,2and3 voltages and the prefilter offset were so far off nominal that the autotune could not adjust.

Thanks!

I noticed that the 3rd lens has a voltage of 40V, which is really to high for the normal 7V. Lens is working overtime to focus into the quadropole.

Technician suggested I clean again, maybe did not sonicate that well last week. Looks to him like it's just dirty. So here I go again :(
SM
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry