Chromatograms
Posted: Wed Mar 03, 2010 3:58 pm
by sribala
When I am analysising the samples using GC using Capillary Columns which Peak should come first i.e Formic or Acetic acid i.e Based on Molecular weight or not. But the Stanadard supplier has given the chromotogram as first Acetic acid Follwed by Formic acid. Wheras I am under the impression that it should be in reverse order. Please clarify
Posted: Wed Mar 03, 2010 9:11 pm
by carl.nott
No clue offhand. I would see what type of capillary column you're using and check reference chromatograms for that column (I would imagine they have them on their websites since they have them in the catalogs). I'm assuming your standard has both analytes in it premixed so you can't just run one or the other individually?
Posted: Thu Apr 15, 2010 11:24 am
by Pinal
DELETED BY MODERATOR
Posted: Thu Apr 15, 2010 6:28 pm
by chromatographer1
On a free fatty acid phase column, FFAP for short, the elution is acetic first followed by formic acid. the acids then follow in carbon number order.
Isn't chemistry marvelous?
Rodney George
consultant