Calculation of S/N ratio by Waters Empower software
Posted: Fri Feb 26, 2010 8:29 am
by chhubert
Hi all,
I want to calculate the signal-to-noise ratio of a chromatographic peak using Empower software. Is it possible to do it in this software?
Hubert
Posted: Fri Feb 26, 2010 7:35 pm
by DR
Yes, it is. I'm about 99% certain that this has been covered (so please search as I can't remember ho-to off the top of my heat at the moment).
Posted: Sat Feb 27, 2010 12:48 pm
by chhubert
I did search the help for wordings like signal-to-noise, signal, s/n but with no avail. Apparently it didn't cover under the system suitability either. :?:
Posted: Sat Feb 27, 2010 9:30 pm
by vofelyp
Hi
Here is the Waters technical note URL:
http://www.waters.com/waters/support.ht ... &type=TECN
Posted: Sun Feb 28, 2010 12:02 pm
by chhubert
Hi vofelyp
Since your link is an advanced technical note which I couldn't get access to it, wondering if you could download and email to me. My email a/c is chhubert@hotmail.com. Many thanks.
Hubert :lol:
Posted: Wed Mar 03, 2010 11:49 pm
by Dividination
Yes it is. If you have the Suitability module it is right in the center of the Suitability tab in the processing method. Be carefully about using the Baseline Noise value though (doesn't work well), and also carful using Detector Noise (RMS noise). The Noise range settings are in the Noise and drift tab for the processing method. Note the Suitability tab even has the option to calculate EP, oh wait now I mean USP noise.
Of course if you don't have Suitability you can make a custom calc to divide peak height by noise (or is that 2x the peak height). Also beware of the scaling factor difference between height and noise, but that value is also available for a custom calc.
I have seen signal to noise not available in Empower 1
Posted: Fri Apr 09, 2010 2:22 pm
by EmpowerGuru
It works well in Empower 2
Re: Calculation of S/N ratio by Waters Empower software
Posted: Mon Jul 15, 2013 6:47 pm
by jayroseville58
could anyone show the detailed procedure on doing this with the software. I'm just confused because the limit being set for the sensitivity solution i had should be NLT 10 but i'm getting only 0.8.
Re: I have seen signal to noise not available in Empower 1
Posted: Thu Sep 25, 2014 2:50 pm
by jayroseville58
Re:
Posted: Thu Sep 25, 2014 2:51 pm
by jayroseville58
Hi
can you email it to me as well. jayroseville58@yahoo.com thanks
Re:
Posted: Thu Sep 25, 2014 2:53 pm
by jayroseville58
can you send me your detailed explanation on this part? thanks. here is my email ad. jayroseville58@yahoo.com
Posted: Tue Sep 30, 2014 9:52 am
by FranckV
You can also create a custom Field using this formula :
[Height]*Scale to µV/(Peak to Peak Noise)