5:5:2 what does it mean?
Posted: Wed Feb 24, 2010 10:44 pm
by Ronaldo
( 5 : 5 :2 v/v)
i think it 5ml from the first solvent , and 5 ml from another solvent , and 2 ml
but what if i have to make up to 100 ml with these three solvent using the same percent ( 5 : 5 :2 v/v)
thanks
Posted: Thu Feb 25, 2010 4:20 am
by bisnettrj2
If it were me, I would do the following:
Fill one 50 mL volumetric with solvent A
Fill one 50 mL volumetric with solvent B
Fill one 20 mL volumetric with solvent C
Combine all in a beaker, allow to equilibrate (in case of any exothermic/endothermic mixing reactions), then fill a 100 mL volumetric with the mixture. That way, you keep your 5:5:2 ratio without having to make any complicated measurements of volume.
If you need to make more, or will need more in the future, you could go 250:250:100 and combine to give 600 mL of a 5:5:2 mixture.
Posted: Thu Feb 25, 2010 12:27 pm
by Ronaldo
Thank u very much bisnettrj2 for your way of prepare it but the combine mixture will be ( 120 ml ) how can i put it in 100 ml flask ?
any way i like your method
Posted: Thu Feb 25, 2010 12:40 pm
by danko
Ronaldo,
If you absolutely insist on preparing 100 mL of the mixture, then you’ll need to use the following volumes:
41.67 mL : 41.67 mL : 16.67 mL
Does it help?
Posted: Thu Feb 25, 2010 1:47 pm
by lmh
bisnettrj2 suggested making up a slight excess because if you mix solvents whose volumes add up to exactly 100mL, you may not end up with 100mL. For example, 50mL water + 50mL methanol makes 50%MeOH, but not 100mL! There is a volume change.
Bisnettrj2's method makes exactly the solvent you describe, and relies on you to measure the correct volume.
Posted: Thu Feb 25, 2010 2:19 pm
by bisnettrj2
lmh, danko, thanks for helping me out here...
Posted: Thu Mar 18, 2010 9:45 am
by Ronaldo
speacial thanks to you all