We are using a HP GC 5890 with FID for sampling LPG. We are using a liquid sampling valve. It was workinf fairly good. But the reproducibility is not OK. We are discussing what is the best method for injecting the sample. We are using a regulator at the entrance of the sample. We are using a normal ball valve regulator but I think that the best would be a double stage regulator? What do you think? WE have a regulation valve at the waste end of the valve.
We are sampling the following way:

1. First valve in sampling position
2. open the sampling valve and regulating the pressure to 50 psi
3. open the waste valve and regulating to get the same 50 psi
4. Wait 30 seconds
5. Inject

We have tried many forms of injection and this was the one that gave us the best results.

Do you know where can I find the best method for sampling LPG?
Life is a game, and in a game you can win or loose, but it is still a game.

What is "LPG"?
Jumpshooter

If it's a liquid sample, I wouldn't regulate it. I would plumb right from the cylinder (or quick connect) to the LSV. Then I would add a small section of opaque/clear poly tubing downstream of the LSV and upstream of a needle valve. Then attach the cylinder, close the needle valve, then open the cylinder valve all the way. Now crack the needle valve and use the clear tubing to ensure that you have a liquid full sample loop.

I think he means liquified petroleum gas. Where I work, we have so many three letter abbreviations, that those are called TLAs. Really. Every time I have to go to a #$%#@$$^&&*^$## meeting I hear a zillion that I have no idea what they mean. But many yuppies just assume that everyone knows them all.

Apparently there's a secret spreadsheet that lists them, but it's password protected by a TLA.

Yes LPG means Liquid Petroleum gas and this is the propane gas that we use in the kitchen for coocking.
Life is a game, and in a game you can win or loose, but it is still a game.

If it's a liquid sample, I wouldn't regulate it. I would plumb right from the cylinder (or quick connect) to the LSV. Then I would add a small section of opaque/clear poly tubing downstream of the LSV and upstream of a needle valve. Then attach the cylinder, close the needle valve, then open the cylinder valve all the way. Now crack the needle valve and use the clear tubing to ensure that you have a liquid full sample loop.
Thank you for your answer, we will try it.
Life is a game, and in a game you can win or loose, but it is still a game.
6 posts Page 1 of 1

Who is online

In total there are 99 users online :: 2 registered, 0 hidden and 97 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 97 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry