Does anyone know if ethane sulfonic acid is LCMS compatible as an ion pair reagent? If not could anyone make a suggestion as to a good ion pair reagent for LCMS/UV use at 205 nm? Thank you for you help

Will you us in positive or negative mode?

The instrument gets used in both modes, but mainly positive. This analysis will be preformed in positive ion mode.

The longer-chain alkyl sulfonates are widely considered to be non-volatile, but I don't know about the C2 (I suspect it won't work). The usual suspects for MS work are HFBA or TFA, neither of which is a good choice for UV below about 210 nm. I'm not sure you're going to be able to find a dual-purpose negative-charge IP reagent. :cry:

If it were my problem, I'd go with one of the evaporative detectors for the non-MS work.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

All suflonic acids are not-volatile and thus not compatible with the MS...

Thank you all very much for the help
6 posts Page 1 of 1

Who is online

In total there are 46 users online :: 0 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 46 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry