while doing the calibration of Wavelength Accuracy for HPLC we use Caffiene
But what is the reason?

Posted: Mon Feb 22, 2010 6:47 pm
by tom jupille
It's convenient.
It's cheap.
It's soluble in reversed-phase mobile phases.
It has well defined absorption bands at 205 & 273 nm

There are other choices (e.g., anthracene, with an absorbance maximum at 251 nm).

Posted: Sun Mar 07, 2010 5:24 pm
by akashyewale
Thanks Tom for the information & for mentioning the other product