There is not much story to tell actually, tests were done in daily circumstances .
Ok, the only difference between the two columns is the ID. But as we ran 3 different flow rates without significant differences, ( we cannot obtain the optimal flow rate from the e-card), we didn't want to spend time on plotting, not for the 1.7 um colom, not for the 1.5 um. We used the standard flow rates.
It is obvious that peaks eluting early in an isocratic separation are even more susceptible to this broadening,because the volume of early eluting peaks [low k value] is considerably less than for later eluting peaks [larger k value].This observed peak broadening can result in a very significant loss in resolution. In fact,if the extra-column volume of a system is excessive, the performance of any analytical column can be compromised. But our HPLC system has a minimal extra-column volume gives nearly similar plates for both the 4.6 x 50 mm, and 2.1x 50 mm columns.
I can provide some hints on the optimized system for high resolution, it can help others as well without considerabel investment to improve the performances on traditional systems. Crucial point is to minimize the death volume of the system.
Sub 2um tests we usually run on a Shimadzu Hplc 10AT-series, equipped with a low dispersion manual Rheodyne injector 8125 , installed in the columnoven. Tubing from the injector to column 5 cm, ID 0.005 ", from the column to 5uml UV cell same tubing 15 cm. Soon we will have 0.001" tubings, we believe there is still a range for improvement.
Operational temperature 40 C.
Flow rate 0.2 ml/min for the acquity column ID 0.21 x 50 mm(no significant difference with 0.4 or 0.6 ml flowrate)
Flow rate 1 ml / min for the 1.5 um column ID 4.6 x 50 mm.
Software Class-Vp version 7, with the sensitivity: auxiliary range 1, range 1, response 1, aqcuisition frequency 20 ms.
Pressure for both columns is around 4000 PSI.
Acquity Column is a brandnew, wellwashed one.
And indeed, 1.5 um give some more retention, but longer runs with more hydrofobic samples on the Acquity give max. 7000 Plates/5cm.
It wasn't mentioned to bring strange conclusions, only to ask if 7000 TP is the maximum we can achieve on this column in Isocratic mode on conventional system.