External events on Agilent 6890
Posted: Fri Feb 19, 2010 6:50 pm
We have several other GC already configured this way (maybe not exactly the same way), but this one really has me puzzled.
We use an external box to purge the GC valves between runs. These contain a 110v solenoid which is driven by a 110v relay. We use a 24v signal from the GC to turn the purge on/off (hooked to the relay).
When I hook this device up to valve5 using the external events cable, it's on all the time, regardless of the valve setting (on/off). Checked it with a VOM and yes, it's 24V all the time.
The contact closures - events 7 & 8 - work as expected (on and off), but to use them, now I have to add another power supply (likely 12v).
Am I wrong, but I expected that the events 5 & 6 from the external events cable would provide 24v with the valve on and no volts with the valve off.
We use an external box to purge the GC valves between runs. These contain a 110v solenoid which is driven by a 110v relay. We use a 24v signal from the GC to turn the purge on/off (hooked to the relay).
When I hook this device up to valve5 using the external events cable, it's on all the time, regardless of the valve setting (on/off). Checked it with a VOM and yes, it's 24V all the time.
The contact closures - events 7 & 8 - work as expected (on and off), but to use them, now I have to add another power supply (likely 12v).
Am I wrong, but I expected that the events 5 & 6 from the external events cable would provide 24v with the valve on and no volts with the valve off.