how can we identify whether a perticuler compund is fluorascence active or not by looking its structure? suppose if it is flourascence active then how can we select exitation and emission lambda maximum.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
That's more of a spectroscopy than a chromatography questions, and I'm certainly no spectroscopist. There is a good general description of the fluorescence process on Wikipedia:
http://en.wikipedia.org/wiki/Fluorescence_spectroscopy
In practice, almost any molecule that has a double bond will have an energy level that allows absorption of a photon somewhere in the UV (the more conjugated the double-bond system, the lower the energy gap, and the longer the wavelength). Most of these compounds decay back to their ground state via a series of excited vibrational states. In essence, the absorbed energy is dissipated as heat.
If a molecule has large energy gaps between vibrational states, however, the most probable transition may be to decay directly back to the ground electronic state by emitting another photon. Since some of the energy has already been dissipated, that photon will have lower energy (i.e. a longer wavelength) than the one originally absorbed.
In general, fluorescent molecules have fairly rigid, multi-ring structures that lack closely spaced vibrational modes. The excitation wavelength will generally correspond to a UV absorbance maximum; the emission wavelength will be longer.
So: look at the UV spectrum. Set the excitation wavelength of the spectrofluorometer at the absorbance maximum, and scan the emission side out to longer wavelength to see if there is any fluorescence.
http://en.wikipedia.org/wiki/Fluorescence_spectroscopy
In practice, almost any molecule that has a double bond will have an energy level that allows absorption of a photon somewhere in the UV (the more conjugated the double-bond system, the lower the energy gap, and the longer the wavelength). Most of these compounds decay back to their ground state via a series of excited vibrational states. In essence, the absorbed energy is dissipated as heat.
If a molecule has large energy gaps between vibrational states, however, the most probable transition may be to decay directly back to the ground electronic state by emitting another photon. Since some of the energy has already been dissipated, that photon will have lower energy (i.e. a longer wavelength) than the one originally absorbed.
In general, fluorescent molecules have fairly rigid, multi-ring structures that lack closely spaced vibrational modes. The excitation wavelength will generally correspond to a UV absorbance maximum; the emission wavelength will be longer.
So: look at the UV spectrum. Set the excitation wavelength of the spectrofluorometer at the absorbance maximum, and scan the emission side out to longer wavelength to see if there is any fluorescence.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 16
- Joined: Wed Feb 17, 2010 8:53 am
thank you Tom for good sugession to do the analysis of flourascence active compunds by using flourascence detector in hplc.
3 posts
Page 1 of 1
Who is online
In total there are 39 users online :: 1 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 38 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 38 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science