Matrix Problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello,

I have a problem with one HPLC method I use.

Eluent A Water +0.1% TFA
Eluent B ACN +0.1% TFA

C8 column

Gradient from water to ACN.

I inject a sample containing 1,2-dioleoyl-3-trimethylammonium-propane a cationic lipid and 28 times more sugar. The sample is dissolved in ACN/H2O.

The standard is only the lipid in ACN (same concentration).

At the 100 % concentration everything works fine.

If I dilute the standard 1:100 in ACN and the sample 1:100 in ACN/H2O I get the problem that the standard area is only 2/3 of the sample area.

If I adapt the matrix of the standard to the sample (put sugar inside), the area of the standard is as big as the area of the sample.

I can reproduce this effect.

Can anybody explain this effect?

Thank you in advance

A few questions:
1. What's your detector? PDA? ELS?

2. Why is the solvent you're using for your sample different from that for your standard? (I can think of a few reasons but I want to know yours).

3. What do you mean by "At the 100% concentration everything works fine?"

I'm sorry, but I don't think there's quite enough information (for me at least) to begin to give a cogent analysis.

Hi,

1. DAD

2. Because it is not possible to dissolve the sample in ACN and it is also not possible to dissolve the pure lipid in ACN/H2O

3. At the 100% level of my sample the accuracy is ok, the area of the sample and the 100% standard are nearly the same. At the 1% level the areas differ like described.

New information:
I don’t need the sugar in the standard, the water, the sugar is dissolved in, is enough to get similar areas between sample and standard at the 1% level.
But I still don´t understand the reason.

Consider adsorption of the sample to the wall of your vial, and that this adsorption may be modified by the addition of the sugar.
4 posts Page 1 of 1

Who is online

In total there are 7 users online :: 2 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry