Broad peaks
Posted: Thu Feb 18, 2010 1:38 pm
On a double-channel system with Varian 1079-injectors I inject a 250 ppb solution:
1) 5.0 uL on a 25m 0.25mm DB-5 column with DF=0.25 gas flow = 1.3 ml/min.
2) 2.5 uL on a 50m 0.25mm column with two different phases DF=0.1 gas pressure is 22.5 psi
Both columns have a 4m 0.53 mm retention gap connected with a press-fit.
The DB-5 column gives no problems when connected to any of the injectors but the other column gives broad peaks from the first (Hexachlorobutadiene) to the last (Mirex) peak.
Further information:
Solvent is petroleum ether, boiling point 30-60 C.
Column oven temperature is 40 C on injection and a temperature program starts after 1 minute (25 deg/min to 140 and from there on 5 deg/min to 300).
Detection by MS (1) and ECD (2).
ECD temperature is 320 C, make-up gas nitrogen at 25 ml/min.
I tried injections at 270 C splitless, and later PTV at 45 C for 0.1 minute followed by a 200 deg/min ramp to 350 C .
Solvent shouldn't be changed. Endrin is one of the components in the method.
All settings so far give broad peaks. I suppose that I overload the column, but am not sure. I have contacts with our supplier on the problem but all suggestions are welcome.
1) 5.0 uL on a 25m 0.25mm DB-5 column with DF=0.25 gas flow = 1.3 ml/min.
2) 2.5 uL on a 50m 0.25mm column with two different phases DF=0.1 gas pressure is 22.5 psi
Both columns have a 4m 0.53 mm retention gap connected with a press-fit.
The DB-5 column gives no problems when connected to any of the injectors but the other column gives broad peaks from the first (Hexachlorobutadiene) to the last (Mirex) peak.
Further information:
Solvent is petroleum ether, boiling point 30-60 C.
Column oven temperature is 40 C on injection and a temperature program starts after 1 minute (25 deg/min to 140 and from there on 5 deg/min to 300).
Detection by MS (1) and ECD (2).
ECD temperature is 320 C, make-up gas nitrogen at 25 ml/min.
I tried injections at 270 C splitless, and later PTV at 45 C for 0.1 minute followed by a 200 deg/min ramp to 350 C .
Solvent shouldn't be changed. Endrin is one of the components in the method.
All settings so far give broad peaks. I suppose that I overload the column, but am not sure. I have contacts with our supplier on the problem but all suggestions are welcome.