JDiaz » Tue Feb 16, 2010 12:04 pm
Hi!
I just wanted to share with you what I have done to solve my problem of rising pressure, though I haven´t been able to restore my column as I still have splitted peaks.
I found that the source of over -pressure seemed to be in the injector, as pressure increased with every injection, but not when I ran solvent without injections.
I removed the column and the performed the same washing procedure recommended by Whatman but with a restrictor instead of the column.
I had a big peak eluting from the system in the acid step of the wash, which I suspect was a contaminant of some part of the injector.
After complete washing I reinstalled the column and washed it with the same procedure.
I am now injecting water and standards and, for the moment pressure stays much more stable.
I had my ultrapure water device serviced to minimize contamination from the eluents. I hope this help.
Thank you Danko and HFranz for your ideas!
By the way, I´m still waiting for Whatman technical service to answer