Hi all ,

What is the best way to remove carbonates when you measure Carbohydrates on an HPAEC - PAD ?

thx

Re: Carbonates

Posted: Thu Sep 18, 2025 4:52 pm
by eja85
What brand of instrument do you have? Are you concerned with carbonate in the eluent or samples? Are you manually preparing your eluent?

Re: Carbonates

Posted: Thu Oct 02, 2025 12:47 am
by S_G
Assuming you're preparing eluents and don't have an eluent generator. If you're analyzing a lot of sugar samples, I HIGHLY recommend purchasing an eluent generator. No carbonate issues and pump pistons and seals last so much longer!

Here's a list of things I'd do before obtaining a generator.
  • Prepare eluents weekly using high purity 50% NaOH.
  • Never pour from or shake the 50% bottle - only pipette and never closer than 1-2 inches from the bottom of the bottle. When the bottle gets crusty, time to replace.
  • Change the piston/seal wash weekly. Carbonates tend to build up on the pistons and score the pistons and seals.
  • Sparge if you have helium available. If not, a slow flow of nitrogen gas into the bottle to blanket the eluent and keep the CO2 out helps a lot.

Re: Carbonates

Posted: Sat Oct 18, 2025 5:45 am
by liko44
Hi Guys,
Best practice: remove carbonates by converting them to CO₂ and removing the gas, or by trapping bicarbonate on an ion-exchange cartridge — then filter and degas.mpbhulekh
For example, acidify gently (e.g., dilute acetic acid to pH ≈4), immediately sparge or vacuum-degas to remove CO₂, centrifuge and 0.2 µm filter before HPAEC-PAD. Alternatively, pass samples through a small anion-exchange (H⁺-form) cartridge to remove HCO₃⁻. Always validate that the prep doesn’t hydrolyse or alter your carbohydrates.