Initial Baseline noise HPLC - First few minutes
Posted: Mon Aug 11, 2025 1:06 pm
When doing analysis by HPLC, often in the first minute there is a lot of noise on the baseline. What causes it? Something to do with the injection?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
I am talking about the t0 peak. Can you give more details on what is an injector pulse please.Except for the t0 peak (observed when the injector pulse and/or unretained sample actually reaches the detector), no "noise" of any kind should appear on the baseline at all. The system should be fully equilibrated before the analysis starts. You provided almost no information so it sounds like you may be using HPLC for the first time. Please get some local help from an experienced user and review these initial steps.
Basic troubleshooting steps:
Q1: At what time(s) does the noise appear? Is it during the t0 time?
Q2: If you inject a true blank (a blank is just the mobile phase alone, at the same volume w/o the sample?
Q3: If you inject NOTHING, and only move the injector from LOAD to Inject (manually or via an A/I), do you still see the "noise" appear?
For assistance, please provide basic HPLC method details to be your question in context (Basic info equals: flow rate, column dimensions and particle size, measured column void volume, exact mobile phase composition, gradient or isocratic, sample name, injection volume, injection solution used, detection mode with ALL parameters (e.g wavelength & bandwidth etc). Include a chromatogram, with scales shown, illustrating what you describe as "noise".
The pressure change occurs exactly at the start of the analysis (t = 0 min). This change can be observed on the pressure trace (if recorded) at t = 0 min. However, the first baseline disturbance registered by the LC detector is at t near t0 (which is undoubtedly higher than zero). The pressure change caused by the valve switching might cause some disturbance in the composition of the eluent, and it is the zone of changed composition that moves along the column with the speed near L/t0, resulting in peak(s) near t0. However, in general one cannot say how exactly the eluent composition is altered in this case. Depending on the eluent composition, this concentration disturbance may appear very close to the real t0 or before t0 or after t0 on a chromatogram.The change in pressure from the valve position switch (LOAD to INJECT) results in a real pressure change that is measured by the pressure transducer in the flow path between the pump and injector valve.