I have encountered an issue with the micromass Quattro LC. The cone voltage readback remains at zero. All other voltages and heatings (capillary voltage, desolvation, and ion block source temperature) are adjustable. After removing the source block heater, I am able to adjust the cone voltage again, and the cone readback is also correct. I suspect there may be a short circuit or leakage voltage somewhere. Is anyone familiar with this issue?
-
- Posts: 421
- Joined: Wed Jul 18, 2012 6:45 am
I faced a similar issue with Quattro Ultima PT. I think the source block looks similar. This is a description from my memory because we dont't have this instrument any more, so please don't expect it to be 100% accurate:
Open the cover plate of the source, then you should see the wiring to the heaters (blade connectors). In my case, these were bent up a little so they touched the front plate. Try to bend them downwards (= away from the fromt plate) and try again. It might help to put a little bit of glass wool on top of the screws for insulation.
Good luck!
Jörg
Open the cover plate of the source, then you should see the wiring to the heaters (blade connectors). In my case, these were bent up a little so they touched the front plate. Try to bend them downwards (= away from the fromt plate) and try again. It might help to put a little bit of glass wool on top of the screws for insulation.
Good luck!
Jörg
-
- Posts: 3
- Joined: Mon Aug 11, 2025 11:51 am
Thanks for the tip.
I had already removed the cover plate and measured it for short circuits. The resistance measurement showed approx. 13 ohms in the de-nergized state. So it doesn't seem to be due to a defective ion block heater. I will report back when I have further measurements and results. Thank you very much for now.
I had already removed the cover plate and measured it for short circuits. The resistance measurement showed approx. 13 ohms in the de-nergized state. So it doesn't seem to be due to a defective ion block heater. I will report back when I have further measurements and results. Thank you very much for now.
-
- Posts: 3
- Joined: Mon Aug 11, 2025 11:51 am
The solution:
The problem was caused by a defective IonBlock heater. Although the correct resistance of approx. 13 ohms can be measured at the terminals of the IonBlock heater, the insulation (sheathing of the heating wires) must have infinite resistance rather than finite resistance. With the new heater, the cone voltage is working again.
The problem was caused by a defective IonBlock heater. Although the correct resistance of approx. 13 ohms can be measured at the terminals of the IonBlock heater, the insulation (sheathing of the heating wires) must have infinite resistance rather than finite resistance. With the new heater, the cone voltage is working again.
-
- Posts: 421
- Joined: Wed Jul 18, 2012 6:45 am
Glad to see that you found the problem. Thanks for posting the solution, this will help future users when searching the forum.
5 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science