How many GC's can share 1/8" gas lines?

Posted: Wed Jul 30, 2025 2:24 pm
by MSCHemist
I have a question as too how many instruments can share 1/8" helium and FID Air lines until the lines can't support the flow demand and pressure drop is an issue?

We tend to do a lot of 200:1 injections so that puts a fair demand on the helium line at times if multiple instruments are injecting at the same time.

Obviously FID air requires 400ml/min per FID.

We set the regulators at 80 PSI for the gas lines.

Re: How many GC's can share 1/8" gas lines?

Posted: Wed Jul 30, 2025 11:57 pm
by Jake
It will partly depend on how far the tanks are from the GCs, due to line pressure drop. What's the distance?

Re: How many GC's can share 1/8" gas lines?

Posted: Thu Jul 31, 2025 1:47 pm
by Consumer Products Guy
Would you be able to move up to larger lines and then decrease to 1/8" right at the GCs?

Re: How many GC's can share 1/8" gas lines?

Posted: Fri Aug 08, 2025 4:47 pm
by MSCHemist
They are a pretty good distance probably 20' or more and they go inside the wall so replumbing with 1/4" would be tough.

Re: How many GC's can share 1/8" gas lines?

Posted: Sat Aug 09, 2025 1:16 pm
by Peter Apps