Solvent peak tail

Posted: Wed Jul 02, 2025 11:37 am
by vlastimil.miler
Dear and experienced colleagues,
I would like to ask you as a newbie in the world of GC for help: when i inject DCM (sciex FID) solvent peak tails the entire spectrum and does not return near the baseline. I have no idea what is happening, I have never noticed this before. Thank you very much for your help

Vlastimil Miler
Nuclear Research Institute Rez
Husinec, Czech Republic

Re: Solvent peak tail

Posted: Thu Jul 03, 2025 3:31 pm
by DR
More details will get you better responses... (reworded "Rod's Rule").

Re: Solvent peak tail

Posted: Fri Jul 04, 2025 7:41 am
by Peter Apps
Welcome to the forum.

Do you see tailing with other solvents?

Are you doing split or splitless injections?

When did you last exchange or clean your inlet liner?

When did you last put in a new septum?

Peter

Re: Solvent peak tail

Posted: Thu Aug 07, 2025 1:53 pm
by csiquet_flex2000
Once, I had DCM tailing more than usual in splitless mode.
It turned out the column was not installed correctly in the injector.