Buchi automated RP chromatography issues
Posted: Fri Jun 06, 2025 10:25 am
by iippcc
We have a new-ish (2023) Buchi Pure C850 system in our organic chemistry lab that is working _fine_ (ahem, it's already had its pump, collecting arm, touchscreen, and software-related issues, some of which happened halfway through runs causing product loss, and they were quite expensive to fix) but it is generally working OK with normal phase -it does need like 2x the solvent, test tubes and time our ancient Biotage Isolera (2008) takes for the same level of separation of similar samples, but hey ho.
However, I keep finding a problem when doing reverse phase: For RP I start with a few volumes of weak solvent (water) only, then I proceed to increase the ACN gradient. And here's where my problems start:
When purifying similar amounts of the same mixture of 6-7 products and using the same C18 cartridge, my target molecule comes out 95+% pure at around 40% ACN (halfway through the run) in the biotage (the rest of molecules in the mixture elute separately between 25 and 70% ACN).
When separating the same mixture in the Buchi, every product tends to come out mixed with each other during the first 2-3 volumes of 100% water, or maybe a few products come out mixed first, and other few products are eluted as a mixture too in the next tube with like 99-98% water as an eluent.
I am using a 6g C18 Biotage Sfar cartridge for 100-150 mg of crude which should be enough. This is always the same one (I move it from the Buchi to the Biotage and the other way around), so it doesn't seem to be a cartridge-related issue. Increasing the size of the cartridge to 12g or/and using Buchi C18 cartridges hasn't improved things either.
I've had this issue since day one of trying to do RP, and from time to time when my product doesn't come out in the first water only fractions, then it might not come out at all until the end of the run with 100% ACN in which case I will get 2-3 different fractions, with mixtures of a couple products each. Sometimes nothing will come out at all during the run, and then everything will elute together _after_ having reached 100% ACN, ran a few volumes of it, and when equilibrating the cartridge back to 75% ACN for storage.
Something else I've noticed is, I never recover as much sample as I inject (only around, say, 70%).
Only one or two every 10 attempts have resulted in the products being eluted more or less accordingly to the gradient or at least while the gradient is happening (not with water only, not with ACN only, not post-run), but the separation is still quite poor. There's been no changes in the conditions of gradient, cartridges or pressure, and the odd "still bad but at least during the gradient" separation seems to happen at random and very seldom.
Over the last 2 years I've tried different cartridges, pressures (within the limit of course), flowrates, slower and faster gradients... nothing seems to improve the issue.
Having discussed this with the Buchi people who came to install the instrument, engineers, vendors etc, they basically "blamed" it on us for wanting to use the same conditions we use with the Biotage (again, there's no "conditions" when the separation is running with water only, if the products come out _before_ the "conditions" even start!). They even brought their "test mix" (which is for NP separation only, oh well) to show us how everything comes out exactly when it should according to the datasheet. During one of these visits I did an RP run of one of my crude products in front of them and showed them everything eluting at once by the 3rd volume of water (and compared it to how different things eluted differently over a gradient in the Isolera): they said it might be that the liquid injection valve was a bit loose and they tightened it. But the problem persists.
Our interactions with Buchi about this are becoming problematic as our contacts are getting increasingly defensive whenever we mention it, they always hint at us doing something wrong, and actually there's not much use discussing it anymore as none of them are chemists anyway and they don't see anything strange in the retention times we are getting, or in the fact that very fatty-like molecules coelute with hydrophylic ones. They keep insisting that their NP "test mix" is being separated according to the datasheet and that's it.
I never really liked this instrument for many reasons, and it is obvious that even an ancient Biotage system performs much better, but at least I would expect some RP separation from the Buchi, even on the poor side...What might be happening there? Could it be that the Buchi Pure is, for whatever the reason, just not great for reverse phase? Has anyone noticed similar issues with a Buchi Pure instrument on RP? What are your thoughts?
Re: Buchi automated RP chromatography issues
Posted: Fri Jun 06, 2025 3:59 pm
by vmu
Try to:
- read this:
Observations on the Wetting of Reversed-phase HPLC Packings
https://gimitec.com/file/twhplc97.pdf
- wash the column with ACN before the run, then switch to the starting aqueous eluent without stopping the flow and wash the column with a few column volumes of the starting eluent, then, without stopping the flow, inject the sample and start the gradient run;
- change the gradient to make it starting from 20 % ACN instead of 0 % ACN;
- verify that the solvent delivery system works properly during the gradient run (measure the flow rate by using a volumetric cylinder or a flask and a stopwatch).
Posted: Mon Jun 09, 2025 10:36 am
by iippcc
Thanks for your suggestions!
Do you think the advice for the cartridges will work though, considering that it doesn't seem to be a cartridge-related issue. I consistently get very good separations in the biotage but not in the buchi despite using the very same cartridge. Working in the biotage first and in the buchi afterwards or vice versa doesn't seem to change things either.
Still worth a shot of course. And I'm definitely testing the flow. Will keep you posted.
Cheers
Posted: Fri Jun 13, 2025 11:12 am
by iippcc
Try to:
Hi, I tried this and I got the same poor separation with everything coming out too early (but not with the front of the solvent), as usual. When using the same cartridge in the Biotage, I got the good separation I normally get.
Posted: Fri Jun 13, 2025 11:16 am
by iippcc
As for this, the volumes obtained in a period of time seem to be alright. However, I don't know how the system is mixing. I've been thinking of using a mix of solvents in which one of them is colorimetrically measurable, and see how much of it I get at a given point in the run.
This is given me very bad headaches! :roll:
Posted: Sat Jun 14, 2025 6:49 am
by vmu
A dilute solution of acetone in water is usually used for such testing of HPLC solvent delivery systems:
QUALIFICATION OF LIQUID CHROMATOGRAPHY EQUIPMENT
https://www.edqm.eu/documents/52006/128 ... 8491788491
Posted: Sat Jun 14, 2025 12:40 pm
by Hollow
weird but interesting issue.
I'm not familiar with the flash system but coming from the HPLC side.
So what vmu mentioned, it would be good to do some basic QO/PQ measurements on the pump and gradient proportioning unit.
You may also find some tips and protocol on the LC Resources Troubleshooting wizard under: http://www.lcresources.com/resources/TSWiz/
Just adapt the HPLC descriptions to FlashChromatography.
- doing the gradient mixing test with Acetone (about 0.1% in Methanol or Water vs. pure Methanol or Water; just use the same solvent).
Or replace Acetone by a small concentration of Caffeine if you need a non-volatile tracer (e.g. for ELSD)
- when doing the test, replace the cartridge by some plain, narrow tubing (or a back-pressure regulator), just to get some back-pressure on the check valves, but not having a bigger volume and/or could retain the tracer.
- when the tracer is already set up, you may also run the measurement of the gradient delay volume.
(Maybe the impact on the flash systems may be less than in analytical, because of the bigger flow rate and column volumes, but it would not hurt to have an idea of it). Maybe compare it to the one on the Biotage system
- Test the gradient proportioning valve (GPV) with different combinations of the solvent lines.
-- Did you already switch the lines for your problematic separation (e.g. ACN in A and Water in B instead of ACN in B; even change to C and D...).
-Do you have all solvent lines primed with solvent, even the one not in use? I would keep them primed with Methanol 100%, just in case.
It seems to me, "like" the/one proportioning valve doesn't like pure water to open/close, while it is OK with NP solvents. (?)
Will it be better, if you add about 10% of ACN and adjust your gradient?
What about the pressure traces on the two systems? Are they similar?
When going from 100% water (A) to 100% ACN (B), the pressure should first rise until about 90%A and then decreases quite linearly until 100%B.
Do the pressure traces look different when you observe "no", "partial" or "full retention"? -> another indication of a GPV failure, not proper opening of the lines.
Other thoughts:
- How do you load your sample?
Is the elution strength of your sample less or matching the initial composition?
-> maybe the Buchi is more sensitive to such mismatch (but wouldn't explain full retention so probably not the root cause)
- Nevertheless, I would try to run some RP test mixes (have a look at analytical columns CoA, what they use as test mixes) or just experiment with Caffeine and compare the retention times on the two systems.
Maybe start in isocratic, premixed (about 15-20% ACN in Water), then change to isocratic but online mixing, and finally, do some gradient from e.g. 5-50% (?) and compare the outcomes on the two systems.
With this, you would get some basic stuff to discuss when in contact with Buchi.
I think, such basic chromatography would be hard to blame on the user, if the system is still not working well.