APHA/SM 5210B - BOD Troubles
Posted: Tue Apr 22, 2025 4:35 am
I realize that this is not a chromatography question, but as there is no environmental/wet chemistry subforum I thought this was the most appropriate one, feel free to move the thread if there is a better place.
I've been asked to validate and implement method SM 5210B for use in our laboratory for use with industrial wastewater analysis. So far, so good.
We've bought the needed reagents, equipment, glassware, etc' wrote SOP's and started training our analysts.
All the method controls seem normal, except the most imortant one, being the GGA standard, which consistently falls below control limits at approx. 160 mg/L BOD (acceptable limits being 198+-30)
We've tried to cross-check everything:
1)used different sources of water: tap water, purified water, double-purified water from different systems, even asked for water from a friendly neighbor lab (who are iso 17025 accredited.) - same results.
2)used 2 different LOTs of polyseed seed tablets, same results.
3)used 2 different incubators. same results.
4)tried upping the seed solution amount in the standard bottes (Seed DO uptake went up, proportionally standard results went down)
5)tried using different sources of GGA and nutrient reagents (again, courtesy of our neighbor laboratory) same results.
6)argued with the electrode supplier that we cross-referenced all the other interferences and that the electrode must be faulty, supplier sent technician to recalibrate the electrode using zero-point calibration, didn't help, then we changed the sensor cap on the electrode, didn't help
7)we got a replacement electrode from the supplier, got the same results, took the electrode to our neighbor lab, re-read their samples, got comparable readings, so electrode is supposedly in order.
8)got the friends from the neighbor lab to come and prepare several sequences of controls with us, also to see our working procedure, thinking maybe we're doing something wrong, they took one sample serie with them to their lab, our samples got low results, their were normal.
9)rechecked all our calculations thinking maybe there's a trivial math error, calculations seem correct.
Also, all series and trials have been performed by several different analysts, and solutions have been prepared from stock reagents of different lots several times at different points of time.
Again, blank controls are normal <0.2 ppm. Seed control uptake seems normal ~0.6 ppm.
First day readings seem okay ~8.7-9.1 ppm, only the 5 day readings seem too high.
for example
6 ml GGA bottle DO(0) 9.1 ppm, DO(5) 5.3 ppm (Standard 300 ml BOD bottle, 0.6 ppm seed uptake)
BOD = (DO(0)-DO(5)-Seed)*(Vbottle/Vsample)=(9.1-5.3-0.6)*(500/6) = 160 ppm
I have no idea what to do, been joking that maybe we should ask for a ritual cleansing of the lab to remove negative energies, since all the real analytical factors have been rechecked several times...
Jokes aside, rationally I am sure *something* has to be wrong in the process, but neither I, or the other accredited lab staff have been unable to isolate it...
Any assistance would be appreciated.
for example