Hi, very new to GC FID and the Trace gc Ultra device. On powering it up it gives the hardware shutdown screen because it thinks it is at 632oC. Does anyone know how to replace the oven temperature sensor (assuming that’s the problem). Apologies for such a basic question, Jim
First , you may check the sensor if it is really faulty and needs replacement.
You need an ohmmeter to check it.

Unplug the instrument ,remove the right hand side panel of the GC , there is a hex screw at the back side of the panel , loosen it , slide the panel to the back , tilt and remove.

You will see multiple sensor connectors on the electronic board , locate the '' oven '' sensor connector , it has two wires , check the resistance between the wires , it must be around 105 -110 ohms when the oven is cold.

If it is open circuit ( no resistance reading ) or short ( zero ohm reading ) , it is faulty.

For replacement of the sensor , you may google '' Thermo Trace GC service manual '' and you will get the manual , there is full explanation of sensor change procedure on the manual.

Good Luck !
Hi Uzman,
I really appreciate the quick response to my question. I have the manual and check the resistance of it tomorrow and report back. Fingers crossed that is all that is wrong. Many thanks, Jim
Hi Uzman,
I really appreciate the quick response to my question. I have the manual and check the resistance of it tomorrow and report back. Fingers crossed that is all that is wrong. Many thanks, Jim
I checked the resistance of the oven temperature sensor as you suggested and it read 103.6 ohms. On reassembly it now seems to be starting up robustly and reading correctly without error. Looks like it might have been a bit of a dry contact and the investigation with confirmation that the sensor was not faulty sorted the issue. Very many thanks for your help - greatly appreciated. Jim
You wellcome Jim , good to hear that it worked out !
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry