EPA Method 625.1 difficulties.
Posted: Wed Apr 09, 2025 5:48 pm
Simply put, the issue is getting the system to show suitability.
Running the DFTPP mix everything looks good (DFTPP tune passes, Pentachlorophenol and DDT have good response and tailing) but the problem child, Benzidine, is showing a lot of degradation and therefore a lot of tailing.
Here is our setup:
Agilent 9000 GC and 5977B MSD
We run it exactly like this application not from Agilent https://www.agilent.com/cs/library/appl ... 7256EN.pdf
What has been done: Source has been cleaned. new inlet liner, septum, guard chip and column (only had 250 injections on it) and was replaced with one that had only 150 injects. I also cleaned the split vent port and replaced the line.
Any suggestions on what else I can do to get the system suitable?