Carbon Tetrafluoride analysis

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I am trying to set up a GC analysis of carbon tetrafluoride (tetrafluoromethane) but so far the FID (no surprise), ECD and GCMS have very small peak areas for this compound. The last detector at my disposal is a PDHID. Has anybody else tried analyzing for this compound?
TO-15 will get you down to 0.2 ppb easily in full scan.
Pacerlaser,

Yes, a PDHID should work. 14.7 eV ionization potential means argon mode wont work but helium mode should give you low ppm sensitivity depending on loop volume and split ratio if you have one. Should come off column very quickly I should think.

Best regards,
Using the PDHID and a packed column did the trick. Eluted around the 10 minute mark.
Hello, using a helium ion detector (PDHID) gas chromatograph is possible to analyze impurities in high-purity (5N) and above carbon tetrafluoride (CF4): O2 Ar, N2, CO, CO2, CH4 with a detection limit of 5ppb, NF3 with a detection limit of 100ppb. Welcome to our website http://www.lansis17.com/
5 posts Page 1 of 1

Who is online

In total there are 3 users online :: 1 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry