Agilent 1100 multiple detector configuration
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
Just curious on how people have handled configuration of multiple detectors on an 1100 system. Right now I have a DAD, Fluorescence, and a RI detector. The main hplc stack has the DAD at the bottom of the stack, and the Fluorescence and RI are stacked next to the main HPLC system. I'm considering either running in series, DAD --> Fluorescence --> RI, or using a t and running the DAD and Fluorescence in series, and the RI in parallel. Other option is just hooking up the DAD and switching it out with the other 2 detectors as needed. Curious as to how others handle similar setups. Analysis ranges from peptides, beverages, cosmetics, nutritional supplements, etc.
-
- Posts: 112
- Joined: Thu Jan 28, 2016 8:58 pm
In a lab where we had DAD and Fl, we would keep them in the stack in that order. Part of that was because we wouldn't always use the fluorescence detection, but always used DAD/UVD. These would be kept on one stack. With adding RI, or ELSD or any third or fourth detector we would pull the detectors into their own stack right next to the LC Separations stack. This would also allow us to more easily switch between single detectors.
In all honesty, it was overkill. We kept the DAD hooked up first since it was always used for analysis. ELSD was almost never connected in the flow path. RI was sometimes hooked up, but always last. Hope this helps you.
In all honesty, it was overkill. We kept the DAD hooked up first since it was always used for analysis. ELSD was almost never connected in the flow path. RI was sometimes hooked up, but always last. Hope this helps you.
-
- Posts: 525
- Joined: Tue Sep 15, 2015 2:07 pm
In order to make a productive suggestion, we would first have to know how often you require one or more of the detectors when you perform routine analysis? Need will justify what the most logical approach will be. For most users who have multiple detectors available, a full scanning DAD will always be in the first position (The UV/VIS DAD is the most useful detector of all). If regularly used, Florescence (FLD) would be connected IN-SERIES next (be very careful to not plug or obstruct the lines or flow cell of the FLD, or any other detector). Finally, if RID is needed, it should always be connected in series, after the DAD, never in parallel (you would loose so much signal doing that, plus it would add variability and broadening). Ideally, only connect the detector(s) needed for the specific method, in-line, for the analysis method.
As noted, the scanning DAD, if correctly plumbed using short, narrow tubing from the column outlet is the most useful AND should always be first in-line.
While it is possible to install diverter valves to "select" the needed detector, we do not recommend this option. Doing so adds too much complexity for most users (they forget to switch the correct detectors inline AND forget to document it for GLP requirements, failing validation).
Yes, the DAD can be the module at the bottom of the stack, with the other two detectors stacked next to it so the connection can be short). That is how we install them when performing an install for clients (Side-by-Side stack offers the best utility. It allows most users to safely reach the solvent bottles while keeping the connections between modules as short as possible for the best chromatography).
As noted, the scanning DAD, if correctly plumbed using short, narrow tubing from the column outlet is the most useful AND should always be first in-line.
While it is possible to install diverter valves to "select" the needed detector, we do not recommend this option. Doing so adds too much complexity for most users (they forget to switch the correct detectors inline AND forget to document it for GLP requirements, failing validation).
Yes, the DAD can be the module at the bottom of the stack, with the other two detectors stacked next to it so the connection can be short). That is how we install them when performing an install for clients (Side-by-Side stack offers the best utility. It allows most users to safely reach the solvent bottles while keeping the connections between modules as short as possible for the best chromatography).
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I'm in agreement with what Multidimensional posted.
When I was working, we oftentimes changed set-up of our "stacks" to accommodate refractive index, ELSD, conductivity detectors as needed. Most times I used PEEK tubing and fingertight fittings.
Changing detectors should be a very minor task for an experienced chromatographer.
When I was working, we oftentimes changed set-up of our "stacks" to accommodate refractive index, ELSD, conductivity detectors as needed. Most times I used PEEK tubing and fingertight fittings.
Changing detectors should be a very minor task for an experienced chromatographer.
-
- Posts: 6
- Joined: Wed Dec 04, 2024 10:50 pm
It is possible to run that in sequence as long as FLD used at last detector, which means DAD---> RID ---> FLD or RID ---> DAD ---> FLD. It is just because of that the excitation light may destroy the structure of the components/analysts. however, the DAD or RID unlikely damage the components.
Just curious on how people have handled configuration of multiple detectors on an 1100 system. Right now I have a DAD, Fluorescence, and a RI detector. The main hplc stack has the DAD at the bottom of the stack, and the Fluorescence and RI are stacked next to the main HPLC system. I'm considering either running in series, DAD --> Fluorescence --> RI, or using a t and running the DAD and Fluorescence in series, and the RI in parallel. Other option is just hooking up the DAD and switching it out with the other 2 detectors as needed. Curious as to how others handle similar setups. Analysis ranges from peptides, beverages, cosmetics, nutritional supplements, etc.
-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
The pressure rating of the flow cells must be considered at the first place.
FLD has the most sensitive cell according to my knowledge , and RID is the second.
The outlet tubing of those cells are wider than standard tubings , may be 0.5 mm ID , so may cause peak dispersion , if connected at the first position.
Not agree with the excitation light of FLD destroys the structure of components , it is a monochromatic light , but DAD may do it , all the spectrum of UV light passes through the flow cell first, then the monochromator sends the passing UV light to diode array.
FLD has the most sensitive cell according to my knowledge , and RID is the second.
The outlet tubing of those cells are wider than standard tubings , may be 0.5 mm ID , so may cause peak dispersion , if connected at the first position.
Not agree with the excitation light of FLD destroys the structure of components , it is a monochromatic light , but DAD may do it , all the spectrum of UV light passes through the flow cell first, then the monochromator sends the passing UV light to diode array.
6 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science