LC-MS calibration injections and check standard injections
Posted: Sat Mar 01, 2025 1:06 am
We are seeing something puzzling on our LC-MS/MS runs (triple quad). Our batches begin with a calibration curve, then we inject samples, with a check standard (bracketed by blanks) every 10 samples. The calibration curve has excellent linearity and the blanks are indeed blank.
What I find puzzling is that the same sample, in the same vial, gives a certain response during the calibration samples--say, 50,000 cts/sec--and then, for the check-standard injections later in the run, it consistently gives a much lower response--say, 30,000 cts/sec. The same vial is being injected for the calibration and for the checks. It is not steadily drifting, like the first injection gives 50,000, the next gives 45,000, the next 40,000, etc. No, instead, the calibration gives one response, and the checks give a different response, consistently--50,000 cts/sec during calibration, then 30,000, 30,000, 30,000, 30,000 cts/sec for the checks.
We have tried equilibrating the instrument by running ~10 blanks at the beginning of the run, to ensure it is fully warmed up. But somehow, the instrument seems to have a telepathic ability to know whether the vial is being injected for calibration purposes (in which case, it measures 50,000 cts/sec) or for check-standard purposes (in which case, it measures 30,000 cts/sec). I do not believe our instrument has telepathy and is trying to sabotage us, so I do not understand how this is possible.
Am I missing something? Any help whatsoever would be appreciated. Thank you!