In order to make a productive suggestion,
we would first have to know how often you require one or more of the detectors when you perform routine analysis? Need will justify what the most logical approach will be. For most users who have multiple detectors available,
a full scanning DAD will always be in the first position (The UV/VIS DAD is the most useful detector of all). If regularly used, Florescence (FLD) would be connected IN-SERIES next (
be very careful to not plug or obstruct the lines or flow cell of the FLD, or any other detector). Finally, if RID is needed, it should always be connected in series, after the DAD, never in parallel (you would loose so much signal doing that, plus it would add variability and broadening). Ideally, only connect the detector(s) needed for the specific method, in-line, for the analysis method.
If you do not use the FLD or RID often, you may be better off having only the DAD in-line as default, then use a high quality, one piece Fingertight connector to make any needed connections to place either the FLD or RID in series with it, only when needed.
As noted, the scanning DAD, if correctly plumbed using short, narrow tubing from the column outlet is the most useful AND should always be first in-line.
While it is possible to install diverter valves to "select" the needed detector, we do not recommend this option. Doing so adds too much complexity for most users (they forget to switch the correct detectors inline AND forget to document it for GLP requirements, failing validation).
Yes, the DAD can be the module at the bottom of the stack, with the other two detectors stacked next to it so the connection can be short). That is how we install them when performing an install for clients (Side-by-Side stack offers the best utility. It allows most users to safely reach the solvent bottles while keeping the connections between modules as short as possible for the best chromatography).