DETECTION LIMIT ON LCMS

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi,

I recently started working on the LCMS, and I have wondering what is most reliable way to determine the LOD and LOQ for my analyte. Using my calibration curve (standard deviation of the intercept method) gives really high values for the LOD and LOQ. Is there any other acceptable way of doing this apart from S/N? Thank you.
Hi Folarin,

sure there is a different way. Have a look: https://www.agilent.com/cs/library/tech ... 7651EN.pdf
Hi Folarin,

sure there is a different way. Have a look: https://www.agilent.com/cs/library/tech ... 7651EN.pdf
Hi Thomas,

How do I estimate the LOQ using this method? Thank you
Did you read the paper?
You make some injections (7-10) of a small standard and determine the standard deviation. You can then calculate a method detection limit by multiplying the SD with a t-Value (Students t for a certain 1-alpha and n-1 degrees of freedom).
Calculation example is given in the paper.

For a more detailled explanation of the method you could check the EPA document: https://www.epa.gov/sites/default/files ... 3-2016.pdf

The MDL is equivalent to a LOD. To get a LOQ, you can either just multiply by a factor (usually 3.33) or use another statistical criterion, e.g. the lowest spiked concentration that gives an acceptable RSD (e.g. <20%).
Did you read the paper?
You make some injections (7-10) of a small standard and determine the standard deviation. You can then calculate a method detection limit by multiplying the SD with a t-Value (Students t for a certain 1-alpha and n-1 degrees of freedom).
Calculation example is given in the paper.

For a more detailled explanation of the method you could check the EPA document: https://www.epa.gov/sites/default/files ... 3-2016.pdf

The MDL is equivalent to a LOD. To get a LOQ, you can either just multiply by a factor (usually 3.33) or use another statistical criterion, e.g. the lowest spiked concentration that gives an acceptable RSD (e.g. <20%).

Hi Thomas,

I read it 😅.

It didn't really mention anything related to LOQ, and I just wanted to be sure. Thank you for your response.
5 posts Page 1 of 1

Who is online

In total there are 5 users online :: 2 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Semrush [Bot] and 3 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry