Hi All,

Just got a problem with the chemstation software. In the method & run window, all the icons in the data presentation area (just after the column and DAD icons) are grayed. We still can run a test and see the peaks from the online monitor. However the signal data files have not been automatically saved, so we cannot load the files in data analysis. Does anyone know how to fix the problem? DO we have to re-install the software? THX.

p

Exactly what modules do you have? I'm betting you do NOT see a red line on the online monitor window at the time of injection, I've seen that when there needs to be a remote cable attached to the computer to the stack, or from an internal A/D to the stack. If A/D being used, is there a check in the "Save Data" box?

In the menu:

Method/Run Time Checklist/Check the Standard Data analysis.

The Data Acquisition must be checked too if you want data :)

In the menu:

Method/Run Time Checklist/Check the Standard Data analysis.

problem solved. some guy did uncheck it. Thx.
5 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry