Random Large Peak Popping up Unpredictably

Posted: Mon Feb 24, 2025 7:16 pm
by JPENLAND
I'm having a very strange (to me) issue and don't know how to find the solution.

On one of my GCs, only one of them, a very large peak (very tall, skinny, and takes up 99% of the total peak area) is popping up *sometimes*

It's only that one, and it's not every run. I can have it pop up and then run the exact same sample again and it won't be there. Nothing has changed. To make it worse, it doesn't even elute at the same times every time :lol: I've seen it at 4 mins, 25 mins, and 18 mins.... again same GC, column, and method/settings.

All other peaks look normal, and when I delete that one concentrations of other peaks return to normal.

Anyone have a suggestion on how to approach this?

Re: Random Large Peak Popping up Unpredictably

Posted: Mon Feb 24, 2025 8:30 pm
by Peter Apps
It is not a peak, it is a baseline disturbance. But what might be causin git and how to fix it depend son a lot of things that you do not tell us. What GC, what detector, what column, inlet and sample? What temperatures and flows?

Peter

Re: Random Large Peak Popping up Unpredictably

Posted: Mon Feb 24, 2025 9:50 pm
by Consumer Products Guy
Are you experiencing what we term "spiking" or spikes?

Re: Random Large Peak Popping up Unpredictably

Posted: Fri Mar 07, 2025 3:55 am
by AICMM
JPenland,

Are you running an Agilent?

Best regards

AICMM