I'm trying to develop a method for detecting acetone via derivatisation with DNPH using HPLC-DAD.
I have an issue with sample preparation and can't seem to fix it. I always get far more acetone-DNPH in my own derivatised standards than I do with a certified acetone-DNPH standard.
2mg/L acetone-DNPH standard peak area = 0.9938
My own derivatised blank peak area = 11.6328
2mg/L derivatised sample made with certified acetone standard = 15.5618
Blank corrected peak area = 3.9290
My DNPH is DNPH-Phosphoric acid (https://www.sigmaaldrich.com/GB/en/product/sial/42215) diluted 1:250 in acetonitrile and phosphoric acid.
1mL of standard mixed with 4mL of DNPH solution.
Results are consistent regardless of DNPH concentration (tried 1:2500 and 1:25 dilutions), derivatisation time (30 mins - 24 hours) or derivatisation temperature (room temp - 40C at 4 hours).
Method parameters:
60:40 Water:Acetonitrile mobile phase
C18 column (works well with formaldehyde-DNPH)
0.6mL/min
42C column temp
Now I know the blank peak area is big in and of itself, but I'm really struggling as to why even once blank corrected, my self made derivatisation standards are so much higher than the pre-derivatised acetone-DNPH standard and would really appreciate any help with what I'm doing wrong here.