Problems detecting Acetone with DNPH

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi,

I'm trying to develop a method for detecting acetone via derivatisation with DNPH using HPLC-DAD.

I have an issue with sample preparation and can't seem to fix it. I always get far more acetone-DNPH in my own derivatised standards than I do with a certified acetone-DNPH standard.

2mg/L acetone-DNPH standard peak area = 0.9938

My own derivatised blank peak area = 11.6328
2mg/L derivatised sample made with certified acetone standard = 15.5618
Blank corrected peak area = 3.9290

My DNPH is DNPH-Phosphoric acid (https://www.sigmaaldrich.com/GB/en/product/sial/42215) diluted 1:250 in acetonitrile and phosphoric acid.
1mL of standard mixed with 4mL of DNPH solution.

Results are consistent regardless of DNPH concentration (tried 1:2500 and 1:25 dilutions), derivatisation time (30 mins - 24 hours) or derivatisation temperature (room temp - 40C at 4 hours).

Method parameters:
60:40 Water:Acetonitrile mobile phase
C18 column (works well with formaldehyde-DNPH)
0.6mL/min
42C column temp

Now I know the blank peak area is big in and of itself, but I'm really struggling as to why even once blank corrected, my self made derivatisation standards are so much higher than the pre-derivatised acetone-DNPH standard and would really appreciate any help with what I'm doing wrong here.
Hello there!

i am sorry if this may sound silly but it is part of the troubleshooting process. How exactly did you prepare both samples? In other words, is there any chance that you prepared a 2 mg/L acetone mixture and then derivatized it?
Luccas Name

YMC applications specialist (LATAM)
Hello,

I have a similar problem. I have been performing HPLC analysis for the determination of formaldehyde in resin samples by derivatization with DNPH. I first recorded the chromatogram of DNPH and then made working solutions by adding a fixed amount of DNPH against increasing formaldehyde concentration. As the formaldehyde concentration increased, the peak area of the derivative compound formed (derivatization: 30 min at room temperature) also increased, while the peak area of DNPH decreased (which is what I expected). I obtained a calibration graph with a correlation coefficient of 0.9992. However, when I derivatized the formaldehyde in the resin under the same conditions, I could not obtain reproducible chromatograms, the peak area kept increasing continuously, and after 18 hours I obtained a constant peak area. This tells me that I need to do some optimization work for the derivatization process of the sample, I think.
However, when I derivatized the formaldehyde in the resin under the same conditions, I could not obtain reproducible chromatograms, the peak area kept increasing continuously
Could your resin be decomposing to release additional formaldehyde?
4 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry