Hi all,

I'm a liquid chromatography researcher trying to set-up an IAM (immobilized artificial membrane) chromatography method in order to be able to determine CHI IAM values of the unknown pharmaceutical discovery compounds. The issue is that when using a new batch of IAM LC column, basic control compounds tend to have significantly lower retention times and lower CHI IAM values that the ones stated in the literature, while for the acidic ones the opposite was observed (but less pronounced). If I try to analyze the same compounds with an old column, I get better correlation with literature values. This was observed previously and stated in the literature (publication from 2020), but it seems to be more significant with the newer column batches.

Does anyone deal with these kind of measurements? Do you observe the same? How do you use the data that you get if they are not comparable to the old ones?

Any insight would be much appreciated!

KR, Leda
Sounds like a column conditioning issue, looking quickly at the Regis IAM column care guide for example it just says inject until stable retention times are achieved which is a bit vague,

If you inject the same sample on a new column multiple (10-20+) times does the retention time shift at all? Never used IAM but seen some quite extensive column conditioning procedures in my time and this would be my first guess.

Best of luck,
Chromavore
2 posts Page 1 of 1

Who is online

In total there are 381 users online :: 0 registered, 0 hidden and 381 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 381 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry