I was testing the manual SPME procedure using a DVB/CAR/PDMS SPME fiber. However, after just a few injections, I noticed that the white coating was gone, leaving only a brown thin capillary visible. I wonder why this happened and what I did wrong—I don’t understand.
Someone must have pulled the fiber apparatus out of the GC inlet or the sample vial without retracting the fiber into the protective sheath first.

You can easily strip the coating off of the end of the fiber assembly if you don't do everything in the correct order.

One other thing to check is to be sure you're using a hollow GC-inlet liner. If there is glass wool in there, it could strip your coating.

Most likely, someone didn't retract the fiber into the sheath before pulling it through a septum.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry