Internal Standard calculations with Agilent OpenLab CDS

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I am in the process of resurrecting a GC method from a former co-worker. The method uses 3-Pentanol as an IS. In the software, for the processing method, there is a data field for entering a value for the internal standard.
I have a 30,000 mg/mL 3-Pentanol standard. The sample prep calls for 10uL of the IS to be added to 1.0 mL of each standard and each sample. The value I was told to enter in the software is 300. The units for the method are mg/L. By my math, this seems to be incorrect by a factor of 10, unless I'm missing something about how the software calculates the results using the IS, response factors and areas in the calibration.
I suggest that you delete one of these duplicate posts, otherwise the threads are going to get mixed up.
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry