contaminated mobile phase
Posted: Sat Dec 21, 2024 7:39 pm
Someone used an empty 4 liter bottle of HPLC grade water for some sort of waste, the bottle was not labeled in any way to indicate that it was not HPLC water. I used the bottle to make a 0.1%TFA mobile phase. I noticed that my baseline looked horrible, and when running a gradient the pressure was different then what it usually is. I went to make new mobile phase and noticed the 4L bottle of "HPLC Water" had an off smell. I took the pH and it was 4.3, and has a TDS of ~100, its not flammable. My best guess is that it is discarded drinking water samples for ICP-MS analysis, which would have been acidified with 0.1% nitric acid
I attempted to flush the system with IPA at 0.5ml/min over night but still have high pressure, and a nasty baseline. Any suggestions on the best way to flush out the system? I replaced the frit in purge valve, replaced my guard column, and am about to back flush my column.
To make matters worse I was initially making new mobile has due to baseline noise which I suspect may be due to a lamp going out, so now i'm trying to troubleshoot contaminated mobile phase, a column that is dirty from the contaminated mobile phase in addition to a possible dying lamp(6718 hours).
My plan
replace lamp
run lamp diagnostics
backflush column (on different instrument)
flush lines/system of contaminated water
My plan
replace lamp