Interfacing a Waters GBIP HPLC and Micromass LAN MS/MS?

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

5 posts Page 1 of 1
I have been trying to revive an old waters system to no avail. The HPLC and PDA work great, but I cannot get Masslynx to work with them. In Windows XP, they will almost configure, but I get the error that the PCI card cannot be found at address 30 when I try to run the software despite both the HPLC and PDA being recognized during the configuration editor in masslynx. All GBIP addresses are correct. When I tried to start fresh in windows 7, I've ran into even more problems simply trying to run IO libraries. Perhaps I need a different PCI/GBIP card? I was just using the same card to run my 6890GC/5973MSD with no problems at all. I should note, I dont have the mass specs hooked up and am just trying to run masslynx with the HPLC and PDA, I need to figure out the best way to add the MSD into the mix. Perhaps a National Instruments LAN to GBIP adapter? Any and all questions and or advice is gladly welcome.

Heres the rundown of the frankenstein system I am struggling to configure:

OS: Windows 7, or Windows XP (I have tried on both)
PCI Card: 82350B PCI GPIB Card

Software: Masslynx 4.1 SCN 805

HPLC: Waters 2695 (GBIP)

PDA: Waters 2996 (PDA)

MS/MS: Waters Micromass ZQ, Micromass Quattro (I have both, and both are LAN)
You have a special HP card. For Waters control have to use National Instrument IEEE card. The card driver v 2.8 can be found in the Masslynx CD. It is good for Win XP. For Win 7 or Win 10 download the version 17.6 driver.
You have a special HP card. For Waters control have to use National Instrument IEEE card. The card driver v 2.8 can be found in the Masslynx CD. It is good for Win XP. For Win 7 or Win 10 download the version 17.6 driver.

Sounds good, thanks for the reply. Does the LAN cable from the mass spec plug directly into the computer or do I need some sort of switch/hub?

I just ordered a National Instruments PCI-GPIB IEEE 488.2 DAQ Card 183617G-01, it says copyright 1998 but I am hoping it will work.

https://www.ebay.com/itm/395448769695?_ ... R6ih4pfsZA
After trying every possible solution, I figured out the PCI/GPIB card I bought off ebay was the problem and was faulty. I was able to secure another one locally, and alas have the HPLC and PDA talking to the masslynx software. I read in the manual that when using an ancillary detector with masslynx that the signal should be routed using analog cables from the PDA to the MS computer, but since I have no MS, I am wondering if that is the problem. The software recognizes the PDA and HPLC but I cant seem to get it to aquire data. I know the data can be read over GPIB alone since that is how millenium/empower aquire it.

There is a dummy MS tune file, but I believe thats only to be used if the MS is connected but not being used. I have reinstalled the software with no MS selected, but still cant seem to get it to work.

Anyone got any ideas?
Post your e-mail address, I can send you detailed instructions how to set up the PC for MassLynx to communicate with the instrument on LAN.
5 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry