How to use relative response factor in

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
Hi all,

I would like to know how to properly use the Relative Response Factors (RRF) calculated in the following reference (https://analyticalsciencejournals.onlin ... .201500106)

In my case, yield of reactions are calculated using the following formula:
Image

If the product X is the same as the substrate S, no response factor are needed. However if X is another molecule the RRF should be used to correct the yield. This is done as follow:

Image

RRF(IS) does notappear since it would cancel out.
When using it in this manner, I am a bit surprised that when the analyte X is basically double the molar mass of S, the term RFF(s)/RFF(x) is roughly equal to 1.1. I would have expected it to be a larger correction.

Thank you!
What units do you use for RRF?
Peter Apps
Assuming we're talking about UV detection, molar mass has nothing to do with absorbtivity. If it did, would you really need to calculate rrf empirically?

Also, it appears that your equations are incorporating the X/IS term too many times. Internal standard area should be used to divide through your standard and analyte peak areas just once, then you treat the resulting ratios as if they were areas for use in external standard calculations... but that's a different issue.
Thanks,
DR
Image
Sorry forgot to mention that the detector is a FID used after a GC...

Does the X/IS appears too many times? Let me give you an example:
Image
No RRF was used in this example. The yield of S does not need any RRF anyway since they would cancel out. The ration area S/ area IS would be the same as long as the substrate S is not consumed by the reaction. The difference in area of the IS at t=0 and t=1 is due to sampling different volume of the reaction.
To a close approximation for ordinary organic volatiles, an FID is a carbon counter and so it responds per unit mass not per mole.
Peter Apps
I see... then the RRF should indeed be close to 1.

Thank you for your help!
I see... then the RRF should indeed be close to 1.

Thank you for your help!
Yes, an FID is very unselective.
Peter Apps
7 posts Page 1 of 1

Who is online

In total there are 6 users online :: 1 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry