They are two markedly different issues, so I'll go in chronological order.
Long story short, I'm trying to validate 8270E/625.1 - My first calibration looked amazing, but a few weeks later (after a routine PM visit) I have fronting peaks and possible ion suppression affecting linearity. Then, the lab decides to add another column for another GCMS method, so now we're using a 2-hole ferrule, switching between inlets/columns, and I've greatly lost sensitivity and have horrible peak shape.
MS: Agilent 5975C
GC: Agilent 6890A
Carrier gas: Hydrogen
Column: Rxi-SVOCms, 30 m, 0.25 mm ID, 0.50 µm
Split Ratio:10:1
Column flow: 1.2(constant flow)
Inlet Temp: 240
Source Temp: 230
Aux Temp: 280
Problem 1: Calibration Curve
8/21/24 - Installed brand new column, liner, seal & septa. Baked out overnight, then ran 20 or so high standards.
8/22/24 - 8/28/24 : DFTPP target tuned. (69~500K) 442 & 68/70 are high in DFTPP. Fine tuned to get 442 down. 68 and 70 are still right on the edge. I've deduced from the forums that could be a hydrogen carrier gas issue, but if anyone has any tips I'm all ears
8/29/24 : Ran a curve, second source, and mid-range iDOC samples. Curve had excellent low-end sensitivity and RSDs were <10 for all but the trouble compounds. 0.5ug/mL to 100ug/mL. In my 10 years of running 8270 I couldn't dream of a nicer curve.
9/13/24 : Engineer comes in for PM. Cleaned source, replaced pump oil, dusted fans tested filaments, tested tunes (I wasn't here for this)
9/16/24 : Other department switches out my column to run their method
9/23/24 : Switch back to my column, bake overnight
9/24/24 - 9/30/24 : Re-tuned (69~400k) and ran a curve. Slightly lower sensitivity, but horrible peak splitting (front shoulders). The compounds themselves are linear, but my internal standards drop significantly at the high points (basically at where the compounds ppm meets the IS ppm), skewing everything quadratic. I researched about this issue and tried the below:
Lowering EMV (no significant change)
Raising EMV (no significant change)
Various split ratios/flow rates (no significant change)
Alternative column using original GC parameters from august [RTX-5silMS 30mx.25umx.25mm] (variation in sensitivity [earlier eluters respond higher, later respond lower], much more column bleed, RT shifts of course, but no change)
I truly have no idea what caused my perfect calibration to fall out from under me. The only significant difference in both tunes is my august one had the entrance lens a smidge open at 7.5 - usually I've always kept it at 0. Other than intentional variations, all my GC parameters were identical.
Problem 2: Sensitivity with 2 columns into 1 transfer line
While I was trying to fix the above, the lab decided to put in a 2-hole ferrule at the transfer line and set up the GC with each dept's columns, so we don't need to constantly vent the instrument. My column got put in the back inlet, which I assume has never been used ever, so I replaced the seal/liner and went on my merry way.
10/9 - Re-tuned (69~400k) and ran DFTPP. Ions were in (though 68 still fights me), but high-not-failing benzidine tailing. Low responses. Ran a midpoint standard and my peak shape for my early eluters (1,4 dioxane, NDMA etc) is just tail city. Also, my responses are about 1/10th what they were during my previous problems above. Tried a few of my classic tailing fixers:
Cleaned split line (no change)
Tried different liner [Restek Topaz LP, had a Topaz Split prior] (even lower responses - but slightly less tailing)
and one thing I just noticed - Restek recommends 1.2mm inlet seals when using 2 columns. I've been using plain ol' 0.8mm ones. I wonder.
I'm positive this has something to do with having the other column going to the aux. I have it set to the same temperature/flow as my actual column. But I'm no engineer (barely a chemist if I'm honest

Sorry for the novel.
I've got chromatography and parameters if anyone thinks they need more detail.
I appreciate any and all ideas or discussion.