Hi Michael,
I have tried the buttons you speak of, but they don't always work when there is no peak. I can even specify an beginning and ending point, but the software "helps" me by changing my settings.
You are correct and my blanks are clean, flat lines. There is very little to integrate, which is the source of the issue. What I'm after is to get a handle on the noise. Speaking as an analytical chemist, there is always noise in any measurement and the amount of it matters. That's what this is trying to get a handle on: what's the contribution of noise to the signal.
I recognize, though, that I am principally a spectroscopist, not a chromatographer, and I am approaching this problem as such. The confusion in your answer tells me that perhaps I am approaching this in the wrong way. Is there a better/more accepted way of determining limit of detection and/or limit of quantitation in chromatography?
Thank you!
Mark
Maybe the confusion is because I'm thinking in terms of GC/MS which is both a qualitative and a quantitative measurement. You only quantitate a compound
after you've concluded that it's in your sample (or, maybe, that you can't rule out that it's in your sample).
Suppose I have a peak at the retention time for an analyte. The ion associated with the analyte is definitely present, but none of the qualifier ions are present (so I see the molecular ion but none of the fragments). Usually that will happen because a different compound is co-eluting. The concentration may give me a reading if I do integrate the peak, but what I actually conclude is *not detected*. There's nothing to count. No matter how many apples show up, the number of oranges is still zero. The correct integration is no integration: if the software auto-integrates noise, you delete that integration.
If you are using a detector like an FID or ECD, I'm not sure the same thing is true. Can anyone else weigh in on that?
PS: it did come to mind that as a rough estimate you could use peak height on your blanks and see if those tiny numbers even make a difference. Or! If you can't integrate, you could do it the old fashioned way and print the chromatogram then cut out the peaks and weigh them to get an area. In fact I think you should just because it's charming.