Agilent 7890B: Back Inlet Pressure Shutdown

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hello!

I am trying to troubleshoot Back Inlet Pressure Shutdown on my Agilent 7890B.

I removed the septa and do not see a liner. We purchased some liners online but it appears they are too large even though it was the model specified. It was working previously.

The gas pressure looks good on the back of the unit. Trying to figure out where I would like for pressure leaks.
Sorry, having problems following this. Are you new to GC? It was working fine, without a liner? Does the new liner you bought not fit because there IS a liner present, and you're trying to insert another one?

What type of inlet are you using? A standard 7890B S/SL? What liners did you buy, from who, and what were they ones you were previously using? I don't get this.

If there's no liner, there's no liner seal ("O-ring"). Put a correct liner and correct liner seal on the liner.
Hey Jake. Sorry for delay in reply. Yes, I'm pretty new to GC maintenance / troubleshooting.

It's a purged pack inlet liner. We ordered these inlet liners from Agilent via the manual and they appear to be too big.

5080-8732
https://www.agilent.com/store/productDe ... =5080-8732

It was calibrated and verified recently by someone else and got some good runs but then our CRM and LCS started failing and I went to check the septa and liner and did not find a liner.

Picture of injection port after removing weldment
Image
Ah, OK. Thanks. Sorry, lots of experience with 7890B's, but I have no experience at all with that inlet, so can't really help much.
Ah, OK. Thanks. Sorry, lots of experience with 7890B's, but I have no experience at all with that inlet, so can't really help much.
Appreciate you looking into.

Helium is getting to the instrument. We think it might be a bad column. I have someone here taking a look at.
5 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry