Large Volume Injection - Solvent Vent Conditions
Posted: Tue Oct 01, 2024 7:37 pm
I've been having a bit of play recently with 'larger' volume injections to try to get my head around all of the functionality of a new instrument. I have been using Agilent's solvent vent calculator to help build some basic vent parameters and these have been 'not unsuccessful' so far but I would be grateful for a bit of advice from folk with a bit more experience.
The solvent vent calculator suggests initial temperatures which are pretty much unobtainable without cryogenic cooling (which we don't have), so I initially used a start temperature of 40C which was OK, but leads to ridiculously long inlet cooling times (~30-40 minutes to get back down to 40C). Whilst I did attempt to address this by cooling the inlet during the run itself, but when the oven is force cooled at the end of a run the inlet heats up again as the pressure of the fans presumably forces hot air into the inlet, so it's a bit of a waste of time!
To address this, I selected a 50C start temperature and this has given more reasonable cycle times and some solid results for mid to late eluting analytes giving ~91-155% of expected peak area. However, my early eluters (BP of ~200C) have proved a bit more troublesome and it's clear that I'm losing analyte along with my solvent. I know that dropping initial inlet temperature is the obvious solution, but wondered if there might be an alternative way of achieving similar outcomes. I can reasonably change vent flow, vent pressure and vent time and wondered what you would suggest?
Kind Regards
TD2
Kind Regards
TD2